Message Date From Subject 5398 2010-11-08 13:01:50 Robert Cerreto KA7EXM Power Meter Reality Check I need a reality check,
I am building a derivation of Roger Hayward's power meter. Roger used LM324 OP AMPS for some DC conditioning. I do not have any of those on hand, but, I do have some LM2902's . Looking at the data sheets and circuit topology, there does not seem to be any reason why the '2902's cannot be used.
Have I missed something?
73 ES TNKS,
Bob WA!FXT
[Non-text portions of this message have been removed]5400 2010-11-08 21:05:44 ka7exm Re: KA7EXM Power Meter Reality Check Hi Bob.
Yep, the op amps are there just for voltage scaling. The PIC processor reads the voltages once every 100mS or so (if I remember correctly). There's no fancy time scaling or high speed sampling, etc. So any pin-compatible voltage scaling would work just fine.
73.
Roger